    The good doctor needs to review his Health Physics in some more detail. Annual dose limit for public to skin is 50mSv. He's also forgotten that beam penetration does not equate to image penetration.

    http://blogs.sfweekly.com/thesnitch/...rsy_ucsf_r.php

    We'll start with what the clinicians have to say. Two quotes from Fergie and Ronnie:

    "The conclusions are wrong," Ronald Arenson, professor of radiology, tells SF Weekly of his own ins ution's letter. "People who are totally unrelated to radiation wrote it. ... It was senior faculty at UCSF. They're smart people and well-intended, but their conclusions, I think, were off-base. They don't understand how radiation translates to an actual dose in the human body."
    "The airport scanner thing is totally bogus," says Professor Fergus Coakley, chief of the abdominal imaging section at the UCSF radiology department. When you fly, you're "closer to the sun, there's less shielding from cosmic radiation, so being worried about the scan on the way to the plane ride where you're getting extra radiation is bogus."
    http://www.fda.gov/Radiation-Emittin.../ucm231857.htm

    Regarding the specific “Red Flag” issues raised in the letter:

    First, the letter is correct to note that the TSA-deployed product is a recent model. However, the specification for the x-ray tube for the deployed model is almost identical to the original 1991 product. The stated concern was, “The majority of their energy is delivered to the skin and the underlying tissue.” We agree. However, the concern that “the dose to the skin may be dangerously high” is not supported. The recommended limit for annual dose to the skin for the general public is 50,000 µSv. The dose to the skin from one screening would be approximately 0.56 µSv when the effective dose for that same screening would be 0.25 µSv. Therefore the dose to skin for the example screening is at least 89,000 times lower than the annual limit.

    Second, radiation safety protection quan ies are stated as ‘effective dose’. NCRP Commentary No. 16 says, “The purpose of effective dose is to place on a common scale the radiation doses: (1) from different types of ionizing radiation that have different biological effectiveness, and (2) in different organs or tissues that have different radiation sensitivities.” Comparing effective doses from different sources is appropriate. The comparison between the effective dose from cosmic ray exposure or a medical diagnostic chest x-ray and the effective dose from a security screening is intended to be a clear means of risk communication.

    The third point relates to a concern “that real independent safety data do not exist.” In fact, independent safety data do exist. Independent measurements have been made on various versions of this product and all results are consistent with the dose specified by the manufacturer. Examples include:

    Sandia National Laboratories, measurements made July 1991. Published as Sandia Report: Evaluation Tests of the SECURE 1000 Scanning System (1992), National Technical Information Service, DE92013773
    FDA, dose measurements re-verified via computational evaluation, September 15, 1998
    N43.17 working group, measurements made at Folsom State Prison on November 15, 1999
    FDA & NIST, Assessment for TSA, July 21, 2006
    Johns Hopkins University Applied Physics Laboratory (JHU APL), Assessment for TSA, October 2009

    Fourth is the concern that “the relevant radiation quan y, the Flux [photons per unit area and time (because this is a scanning device)] has not been characterized.” We disagree that flux is the appropriate quan y. The air kerma (or skin entrance exposure) for one screening can be determined by a direct measurement of the total charge produced in the air contained in an ion chamber during one complete screening when the meter is correctly calibrated. Additionally, measurements to determine the amount of material required to reduce the intensity of the x-ray exposure by half are necessary to convert air kerma (or exposure) to effective dose. These measurements can most practically be made —and indeed have been repeatedly made— at locations where these products are installed and can be made without altering a scanner’s normal operation. These are the same sorts of measurements made to characterize the output of medical x-ray systems.

    Fifth is the assertion that “if the key data (flux-integrated photons per unit values) were available, it would be straightforward to accurately model the dose being deposited in the skin and adjacent tissues using available computer codes [. . .]” In fact, we have done better. FDA and NIST used software called PCXMC to estimate the individual organ doses and to calculate effective dose. This analysis was part of an evaluation performed under contract for TSA. The input information required by the PCXMC program required considerably more information than simply the x-ray flux. Its parameters include 1) the x-ray tube anode angle, 2) anode voltage, 3) total filtration, 4) x-ray field size, 5) location of the field on the body, 6) focus-to-skin distance (FSD), and 7) entrance skin exposure. Every parameter was measured, calculated, or verified by indirect measurement. The modeling results revealed that the dose to the skin is approximately twice the effective dose.
    http://www.tek84.com/downloads/Holt-Letter2010-12-2.pdf

    I am writing to you because of continuing misinformation about the safety of these devices. This is typified by the comments made by Dr. Brenner that you reference in your recent letter to the TSA. Similar comments have been made by professors at UCSF and other universities. In particular, these professors argue that the body scanner radiation only penetrates a few millimeters into the body, resulting in the radiation dose to the skin being far higher than the average dose to the entire person. This line of reasoning has lead to a variety of inaccurate claims:

    - The FDA has seriously miscalculated the radiation safety of these devices.
    - The skin dose is 20 times higher than the effective dose to the entire body.
    - It is inappropriate to use the techniques of medical radiography to regulate
    body scanners, since the x-rays used in medical imaging penetrate deep
    into the body.
    - It is inappropriate to compare background radiation to body scanners, since
    background radiation also penetrates deep into the body.
    - The radiation from body scanners is blocked by clothing, resulting in most
    of the exposure being to the skin of the face and head. This presents an
    elevated risk of skin cancer.

    All of these claims are incorrect, a result of misunderstanding the physics involved. In particular, Dr. Brenner and the other professors have confused Dose Penetration with Imaging Penetration, which are two completely different things. In the attached figures I show measurements taken on a body scanner to help explain this difference.

    Dose Penetration is a measurement of how deeply the energy from the x-ray beam is deposited into the body. A simple way to define and measure this parameter is illustrated in Figure 1. As shown, a radiation meter is placed at the subject location within the body scanner. A certain thickness of plastic,
    simulating overlying body tissue, is placed in front of the meter and the radiation measurement taken. This procedure is then repeated with other thicknesses of plastic. Figure 4 presents the result of such an experiment, which I conducted a few days ago in preparation for this letter. As shown by the solid curve drawn through the data points, placing 5 mm of plastic in front of the meter reduces the intensity of the x-ray beam from 100% to about 95%. In other words, only about 5% of the total energy of the beam is
    deposited in the first 5 mm of depth into the body.
    Likewise, the curve drops to 50% with a total plastic thickness of about 50 mm. This means that about one-half of the body scanner radiation is deposited within 50 mm of the skin, and one-half is deposited deeper than 50 mm into the body. In comparison, Imaging Penetration describes how deep into the body the acquired image can detect objects. This depends highly on the imaging configuration, that is, where the x-ray source and detectors are placed in relation to the subject. Airport body scanners use backscatter imaging, meaning that they create an image from x-rays that reflect from the first few millimeters of the surface of the body. Only a small fraction of the x-rays that strike the body are used to form this image, with the remaining x-rays being deposited into the body. Figure 2 shows a photograph of a test object used to explain this concept.

    This consists of 16 small squares of copper, placed between 16 sheets of 1.588 mm thick plastic. The upper-left copper square is on the surface of the test object. The copper squares immediately to the right are behind 1, 2, and 3 sheets of plastic respectively. On the second row the copper squares are behind 4, 5, 6 and 7 sheets of plastic, respectively, and so on.

    Figure 3 shows an image of this test object taken on a body scanner. The key feature is that a copper square is less visible in the image as it is placed deeper into the phantom. The lower curve in Figure 4 shows a graph of these data. The upper-left square, having no overlying plastic, is assigned a darkness value of 100%. The upper-right square, covered by 3 sheets of plastic (4.76 mm), is 77% as dark. About 10 mm (6-7 sheets) of plastic is required to reduce the darkness of the copper square to 50%.
    http://www.tsa.gov/assets/pdf/jh_apl_v1.pdf

    This is a 60 page PDF from Johns Hopkins regarding assessment of the scanners. The important synopsis begins on the bottom of page iii (last paragraph above the two bullet points) and goes to page vi
    Last edited by Agloco; 05-20-2011 at 08:57 AM.

  3. #3
    I play pretty, no? TeyshaBlue's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Jun 2006
    Post Count
    13,321
    lol @ bouton's rss feed.

  4. #4
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536
    lol @ bouton's rss feed.
    To your excellent comment, I'll add my excellent content: G F Y

  5. #5
    I play pretty, no? TeyshaBlue's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Jun 2006
    Post Count
    13,321
    lol @ the bot content.

  6. #6
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536
    There have been several cases in big hospitals and radiation clinics were the machines were badly regulated by certified radiation technicians, with the resulting being that 100s or 1000s of paitients were over-radiated.

    Looking at the low-wage losers TSA losers and probably the equally low-wage techs who tweak the machines (but probably not too much, gotta save preventative maintenance costs), what are the chances of over-radiation, esp at the high-volumes of "patients" going through the machines, with women in the early states of (yet unknown) pregnancy, with young children?

    Of course, the pretext is that we are being radiated and patted down "for our own good". They can all Go Themselves.

  7. #7
    I play pretty, no? TeyshaBlue's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Jun 2006
    Post Count
    13,321
    Looking at the low-wage losers TSA losers and probably the equally low-wage techs who tweak the machines (but probably not too much, gotta save preventative maintenance costs), what are the chances of over-radiation, esp at the high-volumes of "patients" going through the machines, with women in the early states of (yet unknown) pregnancy, with young children?
    And on cue, the strawman.

  8. #8
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536
    Gfy

  9. #9
    I play pretty, no? TeyshaBlue's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Jun 2006
    Post Count
    13,321

  10. #10
    I am that guy RandomGuy's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Jun 2005
    Post Count
    51,121
    There have been several cases in big hospitals and radiation clinics were the machines were badly regulated by certified radiation technicians, with the resulting being that 100s or 1000s of paitients were over-radiated.

    Looking at the low-wage losers TSA losers and probably the equally low-wage techs who tweak the machines (but probably not too much, gotta save preventative maintenance costs), what are the chances of over-radiation, esp at the high-volumes of "patients" going through the machines, with women in the early states of (yet unknown) pregnancy, with young children?

    Of course, the pretext is that we are being radiated and patted down "for our own good". They can all Go Themselves.
    Given that the machines are pretty low intensity it seems that even a mis-calibrated or malfunctioning machine would burn out before it subjects anyone to anything dangerous. This is my laymans take on it.

  11. #11
    Cogito Ergo Sum LnGrrrR's Avatar
    My Team
    Boston Celtics
    Join Date
    Oct 2008
    Post Count
    22,399
    Thanks for the info Agloco. I think the ethical/liberty concerns over full-body scanners far outweigh any medical argument.

  12. #12
    selbstverständlich Agloco's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Jan 2007
    Post Count
    9,019
    Thanks for the info Agloco. I think the ethical/liberty concerns over full-body scanners far outweigh any medical argument.
    Agreed. Taking the medical/safety part out of the equation ensures that the disagreement is over the correct principle.

    There have been several cases in big hospitals and radiation clinics were the machines were badly regulated by certified radiation technicians, with the resulting being that 100s or 1000s of paitients were over-radiated.

    Looking at the low-wage losers TSA losers and probably the equally low-wage techs who tweak the machines (but probably not too much, gotta save preventative maintenance costs), what are the chances of over-radiation, esp at the high-volumes of "patients" going through the machines, with women in the early states of (yet unknown) pregnancy, with young children?

    Of course, the pretext is that we are being radiated and patted down "for our own good". They can all Go Themselves.
    Much lower actually given that the exposure time is necessarily less due to the high throughput.

    Read what I posted above. The machine is fundamentally limited as to the intensity of the x-ray fluence and energies that can be produced. If those energies or exposure times were to climb to an unacceptable level, then image quality would be substantially degraded. Your TSA monkeys are in tune with that one parameter more than anything.

    Given that the machines are pretty low intensity it seems that even a mis-calibrated or malfunctioning machine would burn out before it subjects anyone to anything dangerous. This is my laymans take on it.
    RG gives a much clearer and easier to digest version than I can.

    Indeed, the anode would melt long before any harmful exposure levels could be reached.
    Last edited by Agloco; 05-21-2011 at 09:24 AM.

  13. #13
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536

  14. #14
    🏆🏆🏆🏆🏆 ElNono's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Apr 2007
    Post Count
    153,473
    Sandia National Laboratories, measurements made July 1991. Published as Sandia Report: Evaluation Tests of the SECURE 1000 Scanning System (1992), National Technical Information Service, DE92013773
    FDA, dose measurements re-verified via computational evaluation, September 15, 1998
    N43.17 working group, measurements made at Folsom State Prison on November 15, 1999
    FDA & NIST, Assessment for TSA, July 21, 2006
    Johns Hopkins University Applied Physics Laboratory (JHU APL), Assessment for TSA, October 2009
    The problem, as I understand it from the article(s), is that a lot of these 'independent' findings have been redacted, which makes comparing methodology and results pretty much a problem, not to mention it bas izes the scientific method of being able to verify the claims.

  15. #15
    Alleged Michigander ChumpDumper's Avatar
    My Team
    San Antonio Spurs
    Join Date
    May 2003
    Post Count
    154,406

  16. #16
    dangerous floater Winehole23's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Nov 2008
    Post Count
    113,802

  17. #17
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536
    EU has much stronger privacy standards (eg, google was forced to allow wireless router owners to opt out of google's invasive harvesting of wireless routers and their locations), bans lots of chemicals that cause cancer or endocrine disruption but are allowed in USA.

    TSA airport security is nothing but bull theatre. Any casual observer of TSA gate staff shows TSA is clearly an employer of last resort.

  18. #18
    selbstverständlich Agloco's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Jan 2007
    Post Count
    9,019
    Well, an RF scanner won't work as well as a backscatter scanner. That much I can guarantee. The physics here is quite basic.

    Also, you're obligated to use active RF scanners for resolutions sake which imparts energy to the target. While not necessarily harmful in terms of cancer induction as far as we know, it's just as nascent as backscatter tech for these uses. So the same problem exists as with the backscatter scanners: Lots of folks jumping onto a boat that hasn't been deemed seaworthy yet.

    Just my two cents.

    EU has much stronger privacy standards (eg, google was forced to allow wireless router owners to opt out of google's invasive harvesting of wireless routers and their locations), bans lots of chemicals that cause cancer or endocrine disruption but are allowed in USA.

    TSA airport security is nothing but bull theatre. Any casual observer of TSA gate staff shows TSA is clearly an employer of last resort.
    Well, it's still a large matter of conjecture as to how well the backscatter technology was tested before being implemented. It's quite disturbing to see so many different views on what should really be a simple, streamlined method of safety testing.

    My guess: The RF tech hasn't been properly researched and tested either.
    Last edited by Agloco; 11-16-2011 at 01:31 PM.

  19. #19
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536
    TSA Puts Off Safety Study of X-ray Body Scanners

    The head of the Transportation Security Administration has backed off a public commitment to conduct a new independent study of X-ray body scanners used at airport security lanes around the country.

    Earlier this month, a ProPublica/PBS NewsHour investigation found that the TSA had glossed over research that the X-ray scanners could lead to a small number of cancer cases. The scanners emit low levels of ionizing radiation, which has been shown to damage DNA. In addition, several safety reviewers who initially advised the government on the scanners said they had concerns about the machines being used, as they are today, on millions of airline passengers.

    At a Senate hearing after the story ran, TSA Administrator John Pistole agreed to a request by Sen. Susan Collins, R-Maine, to conduct a new independent study of the health effects of the X-ray scanners, also known as backscatters.


    http://www.propublica.org/article/ts...-body-scanners

  20. #20
    selbstverständlich Agloco's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Jan 2007
    Post Count
    9,019

    Earlier this month, a ProPublica/PBS NewsHour investigation found that the TSA had glossed over research that the X-ray scanners could lead to a small number of cancer cases. The scanners emit low levels of ionizing radiation, which has been shown to damage DNA. In addition, several safety reviewers who initially advised the government on the scanners said they had concerns about the machines being used, as they are today, on millions of airline passengers.
    Just an observation:

    I find it interesting that folks will line up to get a CXR or CT without question, but remain in a tizzy over a modality that imparts around 1% of CXR dose and .01% of a typical CT dose. People need a clear, definitive voice to clarify the risks. It's a god dammed three ring circus tbh.

    Might want to advise people to avoid bananas as well until this investigation takes place.

    The real concern should be for the operators. That's where the risk is.

  21. #21
    Moss is Da Sauce! mouse's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Mar 2003
    Post Count
    26,358
    Trying to get Agloco to take the dangers of radiation seriously is like trying to get the Octomom to use birth control.


  22. #22
    Veteran
    My Team
    San Antonio Spurs
    Join Date
    Mar 2009
    Post Count
    97,536
    "I find it interesting that folks will line up to get a CXR or CT without question, but remain in a tizzy over a modality"

    medical scans are a choice (and the medical predators are infamously guilty of over-prescribing diagnostic imaging, like they are guilty of very treatment in general)

    security scans before are forced, nobody likes to be forced.

  23. #23
    selbstverständlich Agloco's Avatar
    My Team
    San Antonio Spurs
    Join Date
    Jan 2007
    Post Count
    9,019
    Trying to get Agloco to take the dangers of radiation seriously is like trying to get the Octomom to use birth control.

    Ah yes, it was only a matter of time.

    I support your diagram mouse: I firmly believe people should avoid bananas laced with Pu239, I131, and Cs137. Happy?

  24. #24
    Moss is Da Sauce! mouse's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Mar 2003
    Post Count
    26,358
    Might want to advise people to avoid bananas as well until this investigation takes place.

    .

    You and your banana fetish, trying to down play radiation you wont be happy till everyone has Cancer. And to think your salad tossers in here still claim you bring "wisdom" to the forum?


  25. #25
    Moss is Da Sauce! mouse's Avatar
    My Team
    Dallas Mavericks
    Join Date
    Mar 2003
    Post Count
    26,358
    Ah yes, it was only a matter of time.

    I support your diagram mouse: I firmly believe people should avoid bananas laced with Pu239, I131, and Cs137. Happy?
    I don't seek happiness I seek out people who lie to others for personal gain. And expose them before they can do anymore harm to others.


Thread Information

Users Browsing this Thread

There are currently 1 users browsing this thread. (0 members and 1 guests)

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •